Posted: Wed Aug 09, 2006 2:01 pm Post subject: AD16 OMNIBUS Module: Offset and Gain Calibration |
|
|
Greetings all,
Does the AD16 come with each of its AD7722s calibrated, or is that something the user must do?
Unfortunately, I am not at my hardware, and I cannot test this myself. If it is calibrated, then it seems to me that you would want the CoefficientSource property to default to csRom rather than csDefault. Is there a reason one would want to run the AD16 at csDefault?
Many thanks for your help.
Dan |
|
| Back to top |
|
 |
gmiller
Joined: 07 Mar 2006 Posts: 105 Location: Simi Valley, CA.
|
Posted: Wed Aug 09, 2006 3:04 pm Post subject: AD16 OMNIBUS Module: Offset and Gain Calibration |
|
|
The A/D on this Omnibus card can self calibrate if set in the right mode. This calibration is for the internal A/D only. This will not take out external circuitry errors. If the calibration is done the error will be reduced. There is no internal coefficients to be used and this must be done each time the power is turned on. If the program is set to enable this it does take a small amount of time to go through the procedure before data can be taken.
Gary |
|
| Back to top |
|
 |
jhenderson Site Admin
Joined: 07 Mar 2006 Posts: 2254 Location: So. Cal. USA
|
Posted: Wed Aug 09, 2006 4:30 pm Post subject: Re: AD16 OMNIBUS Module: Offset and Gain Calibration |
|
|
| daniel.hartman wrote: | Greetings all,
Does the AD16 come with each of its AD7722s calibrated, or is that something the user must do?
Unfortunately, I am not at my hardware, and I cannot test this myself. If it is calibrated, then it seems to me that you would want the CoefficientSource property to default to csRom rather than csDefault. Is there a reason one would want to run the AD16 at csDefault?
Many thanks for your help.
Dan |
Dan -
We cal all channels of the module at the factory using a custom jig and instrumentation traceable to NIST. We derive the gain and offset coefficients for each channel which optimally correct for the actual in-system gain and offset errors present on each channel.
When we do the cal, we use csDefault, so that unity gain and zero offset are in effect during the derivation of the actual in-system errors. Later, when you select csRom coefficients, you're using the factory-learned values, which should give corrected results for each channel. Each sample which passes through the logic gets corrected using y=mx+b where m is the channel-specific gain and b is the channel-specific offset value learned at the factory. _________________ phpbb1 |
|